31.7.18 To Assess the Functional Outcome of AO/ASIF A3 Distal Femoral Fractures Fixed by Contour Locking Plate as Fixator Internee

Original Article

 

Femoral Fractures by Contour Locking Plate

To Assess the Functional Outcome of AO/ASIF A3 Distal Femoral Fractures Fixed by Contour Locking Plate as Fixator Internee

Muhammad Khalid Syed, Muhammad Iqbal, Attaullah, Shahzad Nadeem Azhar, Tauseef Ahmed Baluch and Sadia Amin

ABSTRACT

Objective: To evaluate the radiological and functional outcome of Association for Osteosynthesis (AO)/ Association for study of Internal Fixation (ASIF) A3 of distal femoral fractures treated with contour locking plates as fixator internee.

Study Design: Descriptive case series study.

Place and Duration of Study: This study was conducted at the Mayo hospital Lahore from February 2018 to August 2019.

Materials and Methods: Patients with close intra articular commutated distal femoral fractures   are admitted and operated with distal femoral contour locking plates by fixator internee principle. Follow up visit were planned as, after 4tHweek, 6thweek and then monthly for 08 months. Radiological union was noted and functional evaluation was done by using the modified Mize outcome criteria and which was graded 1 as excellent, 2 as good, 3 as fair and 4 as failure.

Results: 16 patients with mean age 40.43 years of age were operated. Male patients were 10 (62.5%) while female were 6(37.5%). Patients were assessed with modified Mize criteria and excellent in 7 (43.75%), good in 4, (25%), Fair in 3(18.7%) and failure in 2, (12.5%) patients. No implants failure was noted .NO nonunion was noted. Bone grafting (bone marrow aspiration) was done in 5 cases.

Conclusion: When locking plate is applied as fixator internee, in intra articular comminuted fractures showed an excellent and good results radiological and functional results in AO/ASIFA3 complicated fractures in majority of the patients.

Key Words: AO/ASIF, A3, Fixator internee, Distal femoral fracture, contour locking plate.

Citation of article: Syed MK, Iqbal M, Attaullah, Azhar SN, Baluch TA, Amin S. To Assess the Functional Outcome of AO/ASIF A3 Distal Femoral Fractures Fixed by Contour Locking Plate as Fixator Internee. Med Forum 2020;31(7):74-78.

 

 

INTRODUCTION

From all femur fractures about 6% fractures are distal femur fractures1. These fractures are common in both male and females.  Road traffic accidents, fall from a height   and trauma during playing are the main reason of fractures in young adult’s males. Usually the age limit is between 20 t0 45 years.  These fractures are very common in aged female usually oboe 60 years of age. Their bone is osteoporotic and minor insult leads to the fractures.2,3

 

 

Department of Orthopaedic Surgery, KEMU / Mayo Hospital, Lahore.

 

 

Correspondence: Dr. Muhammad Khalid Syed, Assistant Professor of Orthopaedic Surgery, KEMU / Mayo Hospital, Lahore.

Contact No: 0333-4287008

Email: iqbaldr1069@ gmail.com  

 

 

Received:    January, 2020

Accepted:    March, 2020

Printed:        July, 2020

 

 

 

 

 

 

These fractures are always very complex and problematic for treating surgeon These fractures not only have comminution and but also have intra articular extension.

There are always complex soft tissues injuries around the knee joint. Sometimes quadriceps injury further add fire to fuel the problems4. These fractures are fixed by deferent implants. Every implant has its advantages and disadvantages. Commonly used implants are angle   blade   plate, and dynamic condylar screw. sometimes intramedullary nails are used.   External fixator may be used in some cases and some surgeon advocates early primary arthroplasty.5,6

The commonly used implant are failed due to the nature of bone and injury. The bone at this area is very osteoporotic and there is much comminution with intra articular extension. Some surgeon advocates the use of   retrograde nailing. These nail have the drawbacks of arthrotomy.  Sometimes    nail protrudes   into the knee joint.  Knee stiffness is very common complication of these procedures7. Now a day most commonalty use implant is distal femur locking plate. These plates have many biomechanical advantages.3,8 Due to the many biomechanical advantages the chances of implant failure are minimum at the same time the chances of implant failure in case of   angled blade plate and retrograde nailing are maximum. Locking plates need much energy before they failed9,10. Locking plates can be easily use as   bridge plate The comminution area is bridged and screws are used to fix the fractures.  By this way it acts like an internal fixator. This implant is a gold stander in osteoporotic fractures11.  Due to the biomechanical advantages resistance to plate pull out is equal to the sum of all locking screws resistance in case of locking plate. The most common complication reported in conventional plate and screws is collapse of the fracture in Varus position. This complication is prevented by The multiple fixed angle locking screws in locking plate.12

MATERIALS AND METHODS

This descriptive case series was conducted at Mayo Hospital Lahore from February 2018 to August 2019. All the patients with age between 16 to 50 years of either gender having AO/ASIF A3 supracondylar fractures were included in this study. The patients having pathological fracture polytrauma or evidence of osteomyelitis were excluded from the study. Modified Mize outcome Criteria was used to assess the outcome. Total 16 patients were included in the study, Patients were admitted throughout patients and emergency department. After admission routine investigations are done .X-rays of respective joint in AP and Lateral views are done in all casesc.3D CT scan was done as a routine Skin traction was applied in all patients. After test dose parenteral antibiotics were started. Informed written consent was mandatory in all cases. During history and examination, special attention was given to the soft tissues and neuro vascular status. All fractures were operated on next few days.

Through   lateral   approach, two skin incision marked at the thigh under image intensifier, one distal and second proximal to the fracture site. We open the knee joint and make hole on the extensor expansion without opening the fracture hematoma. We slided the contour plate to proximal end of the fracture in between a plane bone and fascia lata. we used long plates. we made sure that proximal site of the distal screw should be at least 6 to 8 cm away from the fracture site. Under image intensifier, distal    site of plate is fixed by one screw but screw did not lock., Then traction applied by an assistant in somewhat   flexation position of knee. Care was taken to prevent sagging at the fracture site. Fracture geometry, length and rotation checked. Now proximal site of the fracture was opened Here   plate is on or in the  vastus lateralus muscle not on the bone, so there was a gap between the bone and plate .Now skin  incision is  made, fascia incised, palate identified ,two bone holder applies one near  and second away from fracture site ,holding the bone and muscle. By maintaining the traction one screw near the fracture and second away from the fracture applied. Screw is applied as, make a drill hole using sleeve, screw size measured, size is measured crossing plate muscle and bone. While fixing the screw, when the screws cross the bone, one artery forces is placed under the plate to stop the plate to push forward during locking. By this method plate will not sit on the bone but stay in the muscle mass showing a gap between the plate and bone . This is a GOLDEN key of this article. All screws then were locked. At least 5 to 6 screws should be at proximal fracture site and 4 to 6 at distal knee joint In case of open fracture (Now converted in close fracture) about 30 t0 50cc bone marrow aspiration was taken from iliac crest and injected in the fracture site especially on the medial site of the fracture because this site is notorious for delay union or even nonunion. After putting suction drain wounds were closed in layers. Isosmotic knee and hip exercises were started whiten few days post-operatively. Intravenous antibiotic was started. We used available antibiotics in the hospital.   After 24 hours’ drains were removed. Next day    x-rays were done.   Stitches   were removed after 14 days as a policy.  Follow up visits were planed after   4th., 6th, and after 8th weeks. Patients were called after every month. After 8th moth final follows up was completed and study closed.  Radiological and functional outcome    was    evaluated    through    modified Mize outcome criteria17 and graded as excellent, good, fair and failure

Statistical analysis was done by using SPSS version 20.  Categorical variables like   fracture   type   and   gender   was   represented   as frequency    and    percentage    while mean SD was calculated for numerical variables like age.

Modified Mize outcome Criteria

Excellent

All of the following:

loss of flexion <10°;

full extension

no varus, valgus, or rotatory deformity

no pain

perfect joint congruency     

Good 

No more than any 1 of the following:

loss of flexion >20°

loss of extension >10°

Varus deformity >5°

valgus deformity >10°

minimum pain

Fair  

Any 2 of the criteria listed in the previous Category

Failure

Any of the following:

flexion ≤90°

Varus deformity >10°

valgus deformity >15°

 joint in congruency

disabling pain,

irrespective of radiographic appearance

RESULTS

Total 16 patients were included in the study. Male were   10 (62.5%) and female were 6 (37.5%). The age range was from 20 to 50 years of age and Mean age was calculated as 33.63±10.14 years of age. The   average follow up was about 39.49 weeks. Mean time of union was 19.8±4.13weeks. In 10(,62.5%) patients     fractures were involved on right side, while in 6(37.5%) patient’s fractures were on   left side.  5(31.5%) patients have road traffic accidents, 4(25%) patients have a fall and   7(43.75%) patients received fire arm injuries. 9(56.25%)patients were with closed fractures and 7(43.75%) patients were of grade 1 open fracture.  10(62.5%) patients were given spinal anesthesia and 6(37.5%) patients received general anesthesia.   40 minutes was average operating time with a range from 30 to 60 minutes (range 30-60 minutes). 8 days was mean hospital stay time and this time of hospital stay was range from 4 to 15 days. (range 4 to 15 days).   16.3 weeks were average union time of fractures and it range from 12.2 to 24.5 weeks (range 12.2 to 24.5 weeks).    In 5(31.25%) patients bone grafting was done through bone marrow aspiration.

7(43.75%) patients achieved excellent results ,4(25%) patients received good results, 3 (18.75%) patients received fair results and 2 (12.5%)patients received poor results as assessed with modified Mize outcome criteria   In 3(18.75%) patients’ superficial infection was noted and in 1(6.25%)   patient deep infection was noted.   Superficial   infection   was   treated with antibiotics and deep infection was resolved after removal of implant. Implant removal was done after union. Limb shortening was not noted in any case. There was no any mal union and nonunion. All implant was stable through the study so there was no implant failure.  Hospital mortality was not   reported of any case.

In 4(25%) patients’ removal of implant was done. The main reason of implant removal was deep infection in one patient and pain in 2 patients. In one patient implant removal was done on request.   Re-fracture was not reported after implant removal up to3 month. 7 months was mean follow up time and it range from 6 to 8 month (range 6 to 9 months).

DISCUSSION

7(43.75%) patients got excellent result, 4(25%) patients got good result, 3(18.75%)patients got fair and   results achieved in 2(12.5%) patients were failure. Gupta18 reported in 34 (85%) patient’s excellent results,
4(10%) patients’ good results and in 2(5%) patients result was failure. Distal femoral locking plates were used for the treatment of condylar fractures in all patients.

 He reported in 8(32%) patients’ excellent results found, in 8(32%) patients’ results were satisfactory, in 3(12%)patients’ results were unsatisfactory and in 6(24%) patients he reported failure results. Trividi4 used locking plates in his study for treatment of AO type fractures in 25 patients.

Rajaiah3 reported in 7(35%) patient results were excellent s, good outcome in (40%) patients’ results were good fair in 4(20%) results were fair and in 1(5%) patients’ poor outcome was reported, when he used a distal locking plates while treating 20 patients’

Saini19 and colleagues did much work on distal femur fractures with distal locking plates and reported in 21(62%) excellent results were found, in 11(32%) patients satisfactory results were found and in 2(6%)   patients unsatisfactory results were found. Neer scoring system was used in their study for functional assessment while we used Modified Mize outcome    Criteria.

We reported in 3(18.75%) patients’ superficial infection and in 1(6.25%) patients’ deep infection was found.    Patient with Gustilo Anderson type I fracture had deep infection one patient with Type I fracture also got superficial infection. Close fractures were found in other 2 patients with close superficial infection.

Rajaiah3 2(10%) patients were found to be infected, Trivedi4 found infection in his 2(8%) and Poole5 infection to be present in his 2(2%) patients.

In our study we did not find Nonunion in literature Reddy2 2(3.3%) patients got nonunion, Trivedi4 reported nonunion in his 1(4%) patient, Poole5 found infection in his 4(3%) patients and Toro20 reported rate was about 2(16.6%) in his cases. Toro and colleagues found 1(8.3%) patient got implant failure. In their study    Short plates with inadequate number of screws were responsible for implants failure. Ricci21 reported that after working on 335 distal femur fractures which he treated with locking plate and nonunion were found in his 64(19.1%) patients. He documented that many problems from patients are very important in failure rate. Diabetes, smoking, obesity, open fractures and inadequate length of plate are main factors for failure of fixation.

Rodriguez et al22 had the same opinion as Ricci and 9.8% infection rate was reported in total 283 patients with distal femoral fractures. In their nonunion he reported the main factors involved were infection, obesity, open fracture and stainless-steel plates Another pointed out that nonunion was found in his 41% patients with stainless steel plates he used and is about 10 % when he used titanium locking plates.

Henderson8 also pointed out that after reviewing 23 articles and reported that in (0% to 32%) patients’ complication of healing was noted, in 0% to 15% patients delayed union were noted and in 0% to 19% patients nonunion were reported and in 0% to 20% implant failure was found. Poole5. reported very high mortality rate after working 12 months. Relatively higher mean age (72.8 years) was the main reason of this problem. Stainless steel locking plate and titanium plates are not compared in our study. The following factor are not discussed in our study like obesity, diabetes, smoking, plate length and other possible risk factors for infection, non-union and implant failure.  On large scale randomized trials should be conducted to address these issues.

CONCLUSION

We recommend that in comminuted extra articular fractures of distal femur fixator internee with locking plate is best option.

Author’s Contribution:

Concept & Design of Study:

Muhammad Khalid Syed

Drafting:

Muhammad Iqbal, Attaullah

Data Analysis:

Shahzad Nadeem Azhar, Tauseef Ahmed Baluch, Sadia Amin

Revisiting Critically:

Muhammad Khalid Syed, Muhammad Iqbal

Final Approval of version:

Muhammad Khalid Syed

Conflict of Interest: The study has no conflict of interest to declare by any author.

REFERENCES

1.      Martinet O, Corday J, Harder Y, Maier A, Buhler M, Barraud GE.  The epidemiology of fractures of the distal femur. Injury 2000;31(3):62-3.

2.      Reddy JAV, Chary NB. Functional   Outcome   of Distal End of Femur with Locking Compression Plate:  A prospective study.  Ann Int Med Den Res 2016;2(1):362-65.

3.      Rajaiah D, Ramana Y, Srinivas K, Reddy SV.  A study of surgical management of distal femoral fractures by distal femoral locking compression plate osteosynthesis.  J Evid Based Med Health 2016;3(66):3584-3587.

4.      Trivedi NP, Chauhan RH, Padhiyar DR, Gandhi SP. Outcome of fracture of intra articular distal femur treated   with   distal   femur   locking   compression plate. Int J Res Orthop 2015;1(1):
22-27.

5.      Poole EC, Wilson GG, Guthrie HC, Bellringer SF, Freeman F, Guryel E, et al.  Modern distal femoral locking plates allow early weight bearing with a high rate of union and low rate of failure: Five year experience from a United Kingdom major trauma centre. Bone Joint J 2017;99-B:951–7.

6.      Kayali C, Agus H, Turgut A. Is distal femoral locking plate the implant of choice to fix all types of Distal femoral fractures?2018;30(2). Successful results of minimally invasive surgery for    comminuted supracondylar femoral fractures     with LISS: comparative study of multiply injured and isolated femoral fractures. J Orthop Sci 2007; 12:458-465.

7.      Hartin NL, Harris I, Hazratwala K. Retrograde nailing    versus    fixed-angle    blade    plating    for supracondylar femoral fractures: A randomized controlled trial. ANZ J Surg 2006;76(5):290-4.

8.      Henderson CE, Kuhl  LL, Fitzparick  DC,  Marsh  JL. Locking platesfor distal femur fractures: Is there a problem  with  fracture  healing? J Orthop Trauma 2011;25(1):8-14.

9.      Koval KJ, Hoehl JJ, Kummer FJ. Distal femoral fixation:   a   biomechanical   comparison   of   the standard condylar    buttress    plate, a    locked buttress plate and the 95-degree  blade  plate.  J Orthop Trauma 1997; 11:521–524.

10.  Zlowodzki M, Cole PA, Williamson S. Biomechanical evaluation of the less   invasive stabilization   system, angled blade   plate, and retrograde intramedullary nail for internal fixation   of   distal   femur   fractures. J Orthop Trauma 2004;18:494-502.

11.  Greiwe RM, Archdeacon MT. Locking plate technology:  Current concepts.  J Knee Surg 2007; 20(1):50-5.

12.  Heather V, Theresa H, John S. Failure of LCP condylar plate fixation in the distal part of the femur. J Bone Joint Surg 2006;88-A:846-53.

13.  Marsh JL, Slongo TF, Agel J, Broderick JS, Creevey W, DeCoster TA, et al. Fracture and dislocation classification   compendium -2007:   Orthopaedic Trauma Association classification, database and outcomes committee. J Orthop Trauma 2007;21(10): 1-133.

14.  Gustilo RB, Mendoza RM, Williams DN.  Problems in the management   of   type III (severe) open fractures:  A new classification of type III open fractures. J Trauma 1984;24:742-6.

15.  Gupta RK, Rohilla RK, Sangwan K, Singh V, Walia S. Locking plate fixation in distal metaphyseal tibial fractures: Series of 79 patients.  Int Orthop2010;34(8):1285–1290.

16.  Rodriguez-Merchan EC, Gomez-Castresana F. Internal fixation of nonunions. Clin Orthop Relat Res 2004;419:13–20.

17.  Mize RD, Bucholz RW, Grogan DP. Surgical treatment of displaced, comminuted fractures of the distal end of the femur. J Bone Joint Surg Am 1982;64(6):871–9.

18.  Gupta GK, Rani S, Kumar R, Singh B. Analysis of management of supracondylar femur fracture by    locking compression plate. Int J Orthopaedics Sci 2016;2(4): 218-222.

19.  Saini RA, Shah N, Sharma D. Functional outcome of  distal  femoral  fractures  treated  with  DF-LCP [Distal femur locking compression plate]. Int J Orthopaedics Sci 2018;4(1): 439-444.

20.  Toro G, Calabrò G, Toro A, de Sire A, Lolascon G. Locking plate fixation of distal femoral fractures is a challenging technique:  A retrospective review. Clin    Cases    Miner    Bone    Metab 2015; 12(Suppl 1): 55-58.

21.  Ricci WM, Streubel PN, Morshed S, Collinge CA, Nork SE, Gardner MJ. Risk factors for failure of locked  plate  fixation  of  distal  femur  fractures:  An analysis   of  335   cases.   J Orthop Trauma  2014;28(2):83-9.

22.  Rodriguez EK, Boulton C, Weaver  MJ,  Herder  LM, Morgan  JH,  Chacko  AT,  et  al. Predictive  factors  of distal   femoral   fracture   nonunion   after   lateral locked  plating:  A  retrospective  multicenter case-control study of 283 fractures.    Injury 2014;45(3):554-559.